Re: WFI system specs versus use specs

In a message dated 1/28/00 7:23:58 AM Central Standard Time, 
Olsonl@Centocor.com writes:

<< I'd appreciate your comments on the following assertions:
 
 WFI systems run at > 65 or 70 or 75 (depending on the speaker) are
 "sterile".  Thus, the system bioburden limit should be zero.  Any bioburden
 recovered should only be Gram positive environmentals.  The WFI spec of </=
 10 CFU/100 mL represents the process spec only.
  >>

WFI systems typically have very low microbial content because of their 
elevated temperature operating conditions.  However, they are not sterile.  
The bioburden limit needn't be zero, although the vast majority of test 
samples are zeros in any well engineering, installed and maintained system.  
It is important to note though that many firms sample WFI systems in areas 
that do not lend themselves to aseptic sampling.  The observation of counts 
from samples taken in utility areas for example should surprise no one, since 
those are usually uncontrolled areas with no prevision for aseptic sampling.  

Jim Akers