As to the Q's: Follow the guidance that the CDRH has given to industry and inspectors in february: General Principles of Software Validation; Final Guidance for Industry and FDA Staff.
Quote:
3.1.3 IQ/OQ/PQ
For many years, both FDA and regulated industry have attempted to understand and define software validation within the context of process validation terminology. For example, industry documents and other FDA validation guidance sometimes describe user site software validation in terms of installation qualification (IQ), operational qualification (OQ) and performance qualification (PQ). Definitions of these terms and additional information regarding IQ/OQ/PQ may be found in FDA's Guideline on General Principles of Process Validation, dated May 11, 1987, and in FDA's Glossary of Computerized System and Software Development Terminology, dated August 1995.
While IQ/OQ/PQ terminology has served its purpose well and is one of many legitimate ways to organize software validation tasks at the user site, this terminology may not be well understood among many software professionals, and it is not used elsewhere in this document. However, both FDA personnel and device manufacturers need to be aware of these differences in terminology as they ask for and provide information regarding software validation.
end quote.
To test a system, you will need to design tests that prove that the system is suited for its intended use. Some are unit tests, some integration tests, some system tests. You can have "use case" and technical testing, black an white box testing.
Testing can take place in many environments (development, acceptance, operational, automated test system, simulator, ...).
Important is to focus on what you intend to prove. The complexity of conclusive evidence is such that the Q-model does not at  all permit to solve the validation for anything but end-user/very small applications. Professional software development processes are needed, and as stated by the FDA: the Q's are Chinese to advanced software developers.
Paul Herremans
-----Original Message-----
From: Denny J. Walthers [mailto:dwalthrs@walthersassoc.com]
Sent: Saturday, 27 April 2002 16:37
To: PharmWeb Computer Validation
Subject: RE: Retrospective Validation of ...

It is wonderful that the software is running so well and that there
are no errors. However, ever hear of Edsel K. Murphy? It will happen.
The basic difficulty from the very weak position that I occupy is
that software used in medical devices and/or drug, med dev,
biological manufacturing must be validated. That means User
Requirements, System Specification, IQ, OQ, and the PQ is
problematic. For my money, PQ's on software really do not accomplish
anything because the software does not change. ALL OF THE TESTING
should be done in the OQ.
Also, refer to 21CFR Part 11 not just for electronic signatures but
for the electronic records that must be kept as well. AUDIT TRAILS.
Learn about them and put them into everything that you do.Some of us
saw this coming years ago. I personally put audit trails into
software I developed for the company that was working for as far back
as 1983. It just makes good sense.
Spend the time and do it right, you won't have to do it over. That is
from someone who has learned from experience.
Denny J. Walthers