Ambika,

Industry practice is to maintain usage logs for all equipment that must =
be calibrated or otherwise maintained.=20

Yes, there must be a usage log for a pH meter and probe used in =
production. The log must record in GMP fashion what batches the =
equipment was used for. It should also record, or reference another =
recording document, each standardization of the pH meter, including the =
batch identification of each standardization buffer solutions used. =
(Your SOP should guide on the pH's of the buffers used, which generally =
state to bracket the target pH range and be no more than 2 pH units =
apart. Your SOP should also guide on what the acceptable performance of =
the pH probe is, which is usually the slope of the response.) The actual =
performance of the probe during standardization, and that it passed or =
did not, are also to be recorded in the usage log.

Regards,

Frank Bales, Ph.D.=20
 