        Baruch:
        
        If I understand your first situation, you should have a 
        residue limit for AAA and for BBB in CCC. These limits can be 
        calculated independently of the equipment used. The next step 
        is to calculate the limit per surface area based only on those 
        pieces of equipment that can contribute to the residue. The 
        limit for AAA per surface area of equipment a,b,c is based on 
        the limit of AAA in CCC, the batch size, and the equipment 
        surface area. The limit for BBB per surface area of equipment 
        e,f,g is based on the limit of BBB in CCC, the batch size, and 
        the equipment surface area. Lots of people use a MAC 
        calculation. However, I like to use an approach that I 
        presented in a paper published in Pharm Tech in October 1998 
        (it was also published in Pharm Tech Europe in the spring of 
        1999), that separates out the limit in the next product, the 
        limit per surface area, and the limit per analytical sample. 
        Mike Mullen, who co-authored the orginal Lilly article on 
        limits, also likes my approach. You might want to review my 
        article. If I'm correct on the analysis of your first 
        situation, the "another approach" becomes moot.
        
        Note that this calculation of limits of both AAA in CCC and 
        BBB in CCC assumes that there are special issues 
        (interactions??) with the presence of both these residues in 
        CCC. 
        